Shoulders, tangent skims, and defensible integration

FW
f.weissTL29 May 2026#1

Posting this under the heading it deserves: Shoulders, tangent skims, and defensible integration Everything below is what sits behind that.

Setting out an integration problem with the actual figures, because describing it in words has not worked.

Main peak with a trailing shoulder. Valley-to-valley gives 97.7%. A tangent skim gives 98.5%. Both are defensible and they are more than a point apart.

The question I actually have is which convention I should be asking suppliers to report, rather than which one is right.

0 likes 3mo
N
NHuddlestonTL1Member16 May 2026#2

Shoulders would be much easier to settle if anyone reported the denominator. Almost nobody reports the denominator.

3 likes 2mo
MG
m.guerreroTL222 May 2026#3

Gradient delay volume differs between instruments and shifts the whole chromatogram. It is why a transferred method rarely reproduces retention times exactly on a different system.

15 likes 2mo
ST
stopper_traceTL2Member27 May 2026#4
m.guerrero, post #3: Gradient delay volume differs between instruments and shifts the whole chromatogram. It is why a transferred method rarely reproduces retention times exactly on a different system. Go to post

A shoulder on the trailing edge is most often a closely related species rather than an artefact. The way to find out is to change the gradient slope, not to argue about the integration.

Take it as a starting point and not as a specification.

30 likes in reply to #3 2mo
NH
n.hartmannTL231 May 2026#5

Taking post #4 at face value and following it one step further.

Where the shoulders reasoning breaks down for me is the step from the group result to the individual case. That step is almost never argued for.

0 likes 2mo
VS
vial_slopeTL3Regular5 Jun 2026#6

That is consistent with mine, for whatever one more account is worth.

1 like 2mo
VM
v.malinowskiTL29 Jun 2026#7
m.guerrero, post #3: Gradient delay volume differs between instruments and shifts the whole chromatogram. It is why a transferred method rarely reproduces retention times exactly on a different system. Go to post

A chromatogram image at a resolution where you can see peak shape but not baseline detail is worth having and is not the same as the data. Ask for the integration table if the number matters.

10 likes in reply to #3 2mo
N
NLoughranTL313 Jun 2026#8
IB
i.beaulieuTL217 Jun 2026#9

Having read the whole shoulders thread before replying: the question in the first post has not actually been answered yet, and three of us have answered a nearby one instead.

2 likes 1mo
I
IMainwaringTL3Regular20 Jun 2026#10

Where the baseline is drawn is the least documented and most consequential choice in the whole determination. Valley-to-valley, tangent skim and forced-to-zero can span a couple of percentage points on the same chromatogram.

Not the whole picture, but the part of it I can speak to.

9 likes 1mo
R
RidgewayTL3Regular24 Jun 2026#11

A definition problem is doing most of the work in this shoulders discussion. Once the term is pinned down I suspect the disagreement mostly goes away and what is left is small.

3 likes 1mo
IG
i.grimaldiTL228 Jun 2026#12

I would be cautious about generalising from the shoulders example above. It is a good example. It is one example.

0 likes 30d
EF
erratum_fileTL3Regular1 Jul 2026 · edited#13

Where I part company with post #9, and it is a narrow parting.

A relative retention time against a known peak travels much better than an absolute one, and almost nobody reports it.

32 likes 27d
AC
a.cabreraTL25 Jul 2026#14
n.hartmann, post #5: Taking post #4 at face value and following it one step further. Where the shoulders reasoning breaks down for me is the step from the group result to the individual case. That step is almost never argued for. Go to post

Post #13 is the version of this I will quote in future. One addition.

If two laboratories disagree by more than two or three percentage points, work through method, integration, sample handling, whether it was the same lot and the same vial, and whether suitability passed. After all five, a gap needs an explanation.

16 likes in reply to #5 23d
HN
h.nicolaidesTL3Regular8 Jul 2026#15
erratum_file, post #13: Where I part company with post #9, and it is a narrow parting. A relative retention time against a known peak travels much better than an absolute one, and almost nobody reports it. Go to post

Post #13 describes the usual case. This is about the unusual one.

Before anything else: what was the gradient, and at what wavelength? Area percent at different wavelengths is not the same number even on the same sample because different species absorb differently at different wavelengths. With the method stated, I can tell you something useful. Without it, all I can say is that there is one large peak.

1 like in reply to #13 20d
IG
in.guerreroTL211 Jul 2026#16

Shoulders was covered in the wiki last year and the page has a review date on it, which is a better starting point than my memory of a thread.

0 likes 17d
EL
endpoint_lineTL3Regular15 Jul 2026#17

On shoulders, I would rather understate and be corrected upward than overstate and be quoted. That is a house style here and it is a good one.

24 likes 13d
ID
i.dumitruTL218 Jul 2026#18
stopper_trace, post #4: A shoulder on the trailing edge is most often a closely related species rather than an artefact. The way to find out is to change the gradient slope, not to argue about the integration. Take it as a starting point and not as a specification. Go to post

Confirming post #17 from a second method, which matters more than confirming it from a second person.

Method validation is the demonstration that a method can separate the compound from its degradation products and impurities reliably. A method that cannot resolve an impurity from the parent peak will not detect that impurity.

A weak preference rather than a position.

11 likes in reply to #4 10d
DW
diluent_watchTL2Member21 Jul 2026#19
h.nicolaides, post #15: Post #13 describes the usual case. This is about the unusual one. Before anything else: what was the gradient, and at what wavelength? Area percent at different wavelengths is not the same number even on the same sample because different species absorb differently at different wavelengths. With the method stated, I can tell you… Go to post

I read post #17 twice before replying, because I had assumed the opposite.

On integration: where the baseline is drawn matters more than most people realise. On a clean chromatogram with well-resolved peaks the choice is inconsequential. On a chromatogram with a trailing shoulder or a rising baseline it matters. Differences of one to two percentage points between defensible integrations are ordinary.

Noting that I have skin in this question and have tried to discount for it.

0 likes in reply to #15 7d
DN
d.ndiayeTL224 Jul 2026 · edited#20

Adding thanks rather than a view. I do not have a view worth the space.

33 likes 4d
AS
a.salcedoTL3Regular27 Jul 2026#21

Where the baseline is drawn is the least documented and most consequential choice in the whole determination. Valley-to-valley, tangent skim and forced-to-zero can span a couple of percentage points on the same chromatogram.

I would rather be precise about what I do not know than vague about what I do.

1 like 13h

Suggested topics

TopicParticipantsRepliesViewsActivity
Follow-up: Comparing two chromatograms from different laboratories, properly
Comparing two chromatograms from different laboratories, properly — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it…
MMSSNCEFGA+62 67 12k 9h
About the HPLC & UHPLC category
Reversed-phase separations, gradients, columns, integration, and reading a chromatogram properly. This post is a community wiki: any member at trust level 3 or above can edit it, and every edit is recorded…
KBNKBNCPB 4 137 18h
Carryover and the ghost peak from last week's standard
On the subject in the title: Carryover and the ghost peak from last week's standard Working notes rather than a conclusion. What changes if the standard account of carryover is wrong? I ask because I have…
PBFDJMKOHD+80 92 16k 16mo
Where the integrator drew the baseline, and how much it moved the number
Where the integrator drew the baseline, and how much it moved the number — setting out what I have, and where I think it stops being reliable. Posting the method first, because I know what the first three…
FNNPMAVKTV+20 24 1.6k 8mo
Injection volume, overload, and peak distortion — one year on
Injection volume, overload, and peak distortion — one year on Writing it up because I had to work it out twice and would rather nobody else did. Posting a small dataset on Injection volume. It is mine, it is…
DTCNFEGFZV+4 8 1.2k 4mo

Related topics — sharing the tags peak integration, UV & DAD detection, HPLC

TopicParticipantsRepliesViewsActivity
Follow-up: Comparing two chromatograms from different laboratories, properly
Comparing two chromatograms from different laboratories, properly — setting out what I have, and where I think it stops being reliable. I would like to understand what this number means before I repeat it…
MMSSNCEFGA+62 67 12k 9h
Repeatability versus intermediate precision
Repeatability versus intermediate precision — setting out what I have, and where I think it stops being reliable. Trying to establish the boundary conditions on repeatability versus intermediate precision.…
LSKAILWNO+119 134 34k 13mo
Washout with a one-week half-life: the arithmetic — does this still hold?
Washout with a one-week half-life: the arithmetic — does this still hold? — that is the question, and I have not found it answered plainly anywhere I have looked. Washout: setting out the arithmetic in full,…
AFANABANNN+41 45 12k 1d
[2026 update] Reading a chromatogram someone posted without axes
Posting this under the heading it deserves: Reading a chromatogram someone posted without axes Everything below is what sits behind that. Reading a chromatogram, from the point of view of someone who has read…
BOECFA 2 59 2mo
Residual solvents and scavengers: what is looked for — what changed since
Residual solvents and scavengers: what is looked for — what changed since — setting out what I have, and where I think it stops being reliable. What changes if the standard account of Residual solvents and…
ESOFIASLC+37 41 7k 2mo