s.karlsen_rphTL32 Nov 2024#31
MP
m.perrinTL24 Nov 2024#32
erratum_file, post #19: Accuracy: the method measures what you intend to measure. For purity methods, this is tested by spike-and-recover experiments: add a known amount of impurity to a sample and measure whether you recover the added amount. I am not the right person to answer the follow-up to this. Go to post

Confirming post #30 from a second method, which matters more than confirming it from a second person.

A stability-indicating method is one demonstrated to resolve the analyte from its degradation products, usually through forced degradation. Calling a method stability-indicating without that work is a claim rather than a property.

24 likes in reply to #19 21mo
VS
v.szaboTL3Analytical chemist6 Nov 2024#33
citation_index, post #9: Forced degradation studies: deliberately stress the material with acid, base, oxidant, heat, light to generate degradation products and demonstrate that the method can separate them from the parent peak. Acceptance is that the method is stability-indicating. Go to post

System suitability is the ongoing evidence that a validated method is still performing. Validation is done once; suitability is done every run, and it is the one that appears on a certificate.

I would want to see it done twice before believing it once.

7 likes in reply to #9 21mo
OV
o.vukovicTL28 Nov 2024#34

A method transferred between laboratories needs a demonstration that it performs equivalently, not just a document describing it. Transfer is where a great many between-laboratory disagreements originate.

Reading it back, the second half matters more than the first.

1 like 21mo
AF
a.finnegan_rdTL2Dietitian11 Nov 2024#35

Coming back to post #32, because the follow-up matters more than the original answer.

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

0 likes 21mo
VK
v.kirchnerTL213 Nov 2024#36

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

A partial answer, offered because a partial answer beats none.

32 likes 20mo
CL
customs_ledgerTL3Regular15 Nov 2024#37
n.vukovic, post #14: Reporting a result to more decimal places than the method's precision supports is a small dishonesty that appears everywhere. A method with a two per cent relative standard deviation does not support a figure quoted to a hundredth. Go to post

A stability-indicating method is one demonstrated to resolve the analyte from its degradation products, usually through forced degradation. Calling a method stability-indicating without that work is a claim rather than a property.

The uncertainty is in the assumption, not in the calculation.

11 likes in reply to #14 20mo
CV
c.vasquezTL217 Nov 2024#38

Specificity is the first question: does the method separate the analyte from everything reasonably expected to be present? A method that has not been challenged with its own degradation products has not answered it.

The short version is the first sentence; the rest is why.

3 likes 20mo
AA
a.almeidaTL220 Nov 2024#39
c.castellanos, post #2: The most useful single question about a method: what would it fail to detect? Every method has an answer and few documents state it. If it helps: the failure mode here is usually boring rather than dramatic. Go to post

The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number.

Marking that as an opinion rather than a finding.

1 like in reply to #2 20mo
PS
p.silvaTL222 Nov 2024#40
SG
s.grigorescuTL2Member24 Nov 2024#41

This follows post #39 rather than contradicting it.

The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number.

Anyone with a larger sample, please post it.

0 likes 20mo
SL
s.lindqvistTL226 Nov 2024#42

The most useful single question about a method: what would it fail to detect? Every method has an answer and few documents state it.

The reasoning is more useful than the number, which is why I have shown it.

5 likes 20mo
FR
figure_reviewTL2Member28 Nov 2024#43
b.jankowiak, post #5: Understood, and I withdraw the assumption I opened with. Go to post

Documented validation is what separates a number from an opinion expressed numerically. That is the whole reason to ask for the procedure identifier rather than the method name.

14 likes in reply to #5 20mo
GT
g.tammTL21 Dec 2024#44

Understood. Thank you for being specific about the limits of it.

29 likes 20mo
SL
sleep_logTL23 Dec 2024#45
II
i.ilungaTL25 Dec 2024 · edited#46
sleep_log, post #45: Validation is compound-specific and matrix-specific. A method validated for one peptide is a starting point for another and not a validated method for it. Go to post

The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number.

9 likes in reply to #45 20mo
FV
first_vialTL1Member7 Dec 2024#47

Post #46 answers the question as asked. The question underneath it is different.

The distinction between a qualified instrument and a validated method is worth keeping. Both are needed and they fail in different ways.

Where I would look next, rather than where I would stop.

20 likes 20mo
SR
sa.rasmussenTL29 Dec 2024#48

I read post #46 twice before replying, because I had assumed the opposite.

A method that reports the same number for every lot is worth a second look. Real processes vary, and a total absence of variation is a statement about the measurement rather than the process.

That much is documented. The rest is how I have interpreted it.

0 likes 20mo
FN
formulary_notesTL3Regular11 Dec 2024#49

Reading rather than answering, but this is the post I would point somebody at.

5 likes 20mo
AI
an.ibarraTL213 Dec 2024#50
b.jankowiak, post #5: Understood, and I withdraw the assumption I opened with. Go to post

Post #46 put the caveat in the right place and I want to underline it.

Reporting a result to more decimal places than the method's precision supports is a small dishonesty that appears everywhere. A method with a two per cent relative standard deviation does not support a figure quoted to a hundredth.

The conclusion is tentative; the arithmetic underneath it is not.

13 likes in reply to #5 19mo
L
LJankowiakTL3Regular15 Dec 2024#51

Range and working range are different things and a certificate rarely distinguishes them. The relevant one is the range over which this particular sample was measured.

I have written this out at length because the short version keeps being misread.

0 likes 19mo
RN
r.novakTL218 Dec 2024#52
v.szabo, post #33: System suitability is the ongoing evidence that a validated method is still performing. Validation is done once; suitability is done every run, and it is the one that appears on a certificate. I would want to see it done twice before believing it once. Go to post

Building on post #50 rather than restating it.

System suitability is the ongoing evidence that a validated method is still performing. Validation is done once; suitability is done every run, and it is the one that appears on a certificate.

Written quickly, so the reasoning may be tighter than the wording.

27 likes in reply to #33 19mo
OA
o.abrahamsenTL3Regular20 Dec 2024 · edited#53

Robustness testing deliberately varies the parameters most likely to drift — organic percentage, pH, temperature, flow — and shows the result does not. It is the part of validation that predicts whether a method will transfer.

13 likes 19mo
KA
k.adeyemiTL222 Dec 2024#54

Linearity across the working range is a routine demonstration and it constrains how far a result can be extrapolated. A method linear from 80 to 120 per cent of nominal says nothing about a sample at ten per cent.

Two people can read the same figure differently here and both be reasonable.

4 likes 19mo
TS
t.steenkampTL2Member24 Dec 2024#55

Post #53 and I disagree about the size of the effect, not about the direction.

Where a pharmacopoeial monograph exists, a method that follows it inherits a great deal of assurance. Almost nothing discussed here has one.

That is the shape of it. The detail is where I would expect to be corrected.

0 likes 19mo
AW
ai.wikstromTL226 Dec 2024#56
s.grigorescu, post #41: This follows post #39 rather than contradicting it. The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number. Anyone with a larger sample, please post it. Go to post

Thank you for taking the time. That was more work than a reply usually is.

0 likes in reply to #41 19mo
CE
crossover_entryTL3Regular28 Dec 2024#57

An independent laboratory's method being different from the supplier's is not a discrepancy. It becomes one only when the results differ by more than both methods' demonstrated precision.

If anyone has run this properly I would rather read that than my own guess.

19 likes 19mo
BW
br.wikstromTL230 Dec 2024#58

Specificity: the method can distinguish the intended compound from related impurities and degradation products. Tested by comparing results on pure compounds, mixtures of compounds, and degraded samples.

Not a conclusion. A place to stand while looking for one.

8 likes 19mo
GR
gradient_reviewTL2Member1 Jan 2025#59
r.mensah, post #4: Coming back to post #2, because the follow-up matters more than the original answer. The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number. Go to post

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

2 likes in reply to #4 19mo
MM
m.marchettiTL23 Jan 2025#60
an.ibarra, post #50: Post #46 put the caveat in the right place and I want to underline it. Reporting a result to more decimal places than the method's precision supports is a small dishonesty that appears everywhere. A method with a two per cent relative standard deviation does not support a figure quoted to a hundredth. The conclusion is tentative; the… Go to post

Limits of detection and quantitation: LOD is the lowest concentration that produces a signal above background. LOQ is the lowest concentration at which the method meets precision and accuracy acceptance criteria. Both are determined empirically.

0 likes in reply to #50 19mo