d.vukovicTL228 Oct 2024#31
h.karlsen, post #5: Post #4 describes the usual case. This is about the unusual one. A chromatogram image at a resolution where you can see peak shape but not baseline detail is worth having and is not the same as the data. Ask for the integration table if the number matters. That is a description of practice, not a recommendation of it. Go to post

This follows post #30 rather than contradicting it.

System suitability testing: injections run before and during the sample run to establish whether the instrument, column and method were performing when the sample was analysed. If suitability did not pass, the sample results from that run are uninterpretable.

19 likes in reply to #5 21mo
DB
dr_bhattacharyaTL3Physician31 Oct 2024#32
JFitzgibbon, post #12: Post #10 and I disagree about the size of the effect, not about the direction. A shallow gradient resolves close-eluting species that a steep one merges. Two honest laboratories running different gradients can report genuinely different purities on the same vial. Go to post

Worth separating two things that post #28 runs together.

Gradient slope is the single biggest driver of apparent purity differences. A shallower gradient over a longer run resolves more impurities and gives a higher purity figure. A steep gradient produces a tidier-looking chromatogram with fewer visible peaks and gives a lower purity figure. Both are legitimate methods and they will not produce the same number.

I looked this up rather than remembered it, which is the right order.

0 likes in reply to #12 21mo
RM
r.mensaTL24 Nov 2024#33

Column chemistry and particle size: smaller particles (1.7 μm) give better resolution and higher efficiency than larger particles (3.5 μm or 5 μm), at the cost of higher back pressure. Newer methods increasingly use smaller particles.

I am aware this is the third time this month I have made this point.

2 likes 21mo
MD
m.dalgaardTL3Regular8 Nov 2024 · edited#34

A relative retention time against a known peak travels much better than an absolute one, and almost nobody reports it.

8 likes 21mo
AJ
a.jansenTL212 Nov 2024#35
l.chevalier, post #9: Change the wavelength and the proportions change even though the sample has not. That is the reason the wavelength has to be on the document for the number to mean anything. The interesting part of this is the exception, and I do not understand the exception. Go to post

Picking up post #34: that is the part I would want checked first.

A shallow gradient resolves close-eluting species that a steep one merges. Two honest laboratories running different gradients can report genuinely different purities on the same vial.

That holds under the stated conditions and I have stated them.

13 likes in reply to #9 20mo
TD
titration_diaryTL3Regular15 Nov 2024#36

Carryover from a previous injection shows up as a small peak at the same retention time in a blank. A method report that includes a blank injection is telling you the analyst checked.

Take the reasoning and check the arithmetic; I do not always get it right.

28 likes 20mo
IG
i.guerreroTL219 Nov 2024#37

I will take the caveat as seriously as the claim, which is the point of putting it there.

0 likes 20mo
GT
g.tanakaTL3Regular23 Nov 2024#38

I had written a reply contradicting post #36 and deleted it. Here is what survived.

Mobile-phase preparation is a genuine source of between-laboratory variation. Acid concentration and organic modifier both shift retention, and neither is usually specified to the precision that would matter.

5 likes 20mo
BF
b.fonsecaTL226 Nov 2024#39
titration_diary, post #36: Carryover from a previous injection shows up as a small peak at the same retention time in a blank. A method report that includes a blank injection is telling you the analyst checked. Take the reasoning and check the arithmetic; I do not always get it right. Go to post

Method validation is the demonstration that a method can separate the compound from its degradation products and impurities reliably. A method that cannot resolve an impurity from the parent peak will not detect that impurity.

Anyone who has looked at this more carefully, please correct the record.

0 likes in reply to #36 20mo
AA
a.adebayoTL230 Nov 2024#40

Noted, and thank you for writing it out rather than summarising it.

0 likes 20mo

Suggested topics

TopicParticipantsRepliesViewsActivity
Injection volume, overload, and peak distortion
Posting this under the heading it deserves: Injection volume, overload, and peak distortion Everything below is what sits behind that. I have spent a fortnight trying to pin injection volume down and I want…
MICRJNCBHK+43 47 36k 2mo
Area percent versus weight percent: the confusion that causes most arguments
Posting this under the heading it deserves: Area percent versus weight percent: the confusion that causes most arguments Everything below is what sits behind that. Proposing that area percent versus weight…
ADSVCDFNAD+10 14 2.1k 2mo
Why 99.2% and 97.8% on the same lot can both be correct
Why 99.2% and 97.8% on the same lot can both be correct I have a specific reason for asking rather than idle curiosity, and the context is below. I would like to understand what this number means before I…
GHAAJSLCCR+16 20 3.1k 11h
Carryover and the ghost peak from last week's standard — what changed since
Carryover and the ghost peak from last week's standard — what changed since Writing it up because I had to work it out twice and would rather nobody else did. What changes if the standard account of Carryover…
FDLTZYFWAA+98 114 37k 22d
Area percent versus weight percent: the confusion that causes most arguments — what changed since
Area percent versus weight percent: the confusion that causes most arguments — what changed since Writing it up because I had to work it out twice and would rather nobody else did. What changes if the…
VBSADMRLEN+30 34 40k 13mo

Related topics — sharing the tags gradient method, data table, HPLC

TopicParticipantsRepliesViewsActivity
Seven-day rolling means and why weekly readings mislead — one year on
On the subject in the title: Seven-day rolling means and why weekly readings mislead — one year on Working notes rather than a conclusion. Asking about Seven-day rolling directly, because I have read four…
SLMBTPRVKO+78 88 16k 5mo
Aggregated purity results across four services, with caveats — a second dataset
Aggregated purity results across four services, with caveats — a second dataset — setting out what I have, and where I think it stops being reliable. Posting a small dataset on Aggregated purity results…
CWFNHLTHRE+127 151 42k 10mo
Comparing two chromatograms from different laboratories, properly
Posting this under the heading it deserves: Comparing two chromatograms from different laboratories, properly Everything below is what sits behind that. I would like to understand what this number means…
DBKKKFFLA+35 40 22k 7mo
[2026 update] Waist circumference technique and its repeatability
Posting this under the heading it deserves: Waist circumference technique and its repeatability Everything below is what sits behind that. An honest uncertainty about Waist circumference technique rather than…
RZZACNSKG+23 27 5.9k 21h
Deletion sequences: expected mass shifts by residue
Deletion sequences: expected mass shifts by residue — setting out what I have, and where I think it stops being reliable. Deletion sequences keeps being re-asked here and I think that is because the answer is…
SCBBCN 2 59 3mo