Why mass spectrometry cannot tell you about an epimer

D
DOdendaalTL3Regular7 Dec 2024#1

Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below.

I have spent a fortnight trying to pin mass spectrometry down and I want to set out where I have got to, because I suspect the honest answer is duller than the thread this will produce.

What I have: a consistent observation across a small number of cases, collected the same way each time. What I do not have: any controlled comparison, or any reason to think my cases are representative.

The specific question is whether the pattern survives once the obvious confounder is removed. I cannot remove it with what I have.

0 likes 20mo
AA
a.aguirreTL210 Dec 2024 · edited#2

Adducts — sodium and potassium especially — shift the observed mass in predictable increments. A report showing an unexplained plus twenty-two is usually showing you a sodium adduct.

Two people can read the same figure differently here and both be reasonable.

27 likes 20mo
PI
p.iyer_pharmdTL3Pharmacist12 Dec 2024#3

Answering the question the opening post raises rather than the one it answers.

Mass error in parts per million is (observed minus theoretical) divided by theoretical, times a million. On a high-resolution instrument a low single-digit figure is unremarkable and expected.

8 likes 19mo
NO
n.okwuosaTL214 Dec 2024#4
a.aguirre, post #2: Adducts — sodium and potassium especially — shift the observed mass in predictable increments. A report showing an unexplained plus twenty-two is usually showing you a sodium adduct. Two people can read the same figure differently here and both be reasonable. Go to post

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

2 likes in reply to #2 19mo
SS
system_suitabilityTL316 Dec 2024#5
HI
h.iyerTL218 Dec 2024#6

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

It took me longer than it should have to see that.

20 likes 19mo
KO
k.otieno_statsTL3Statistician19 Dec 2024#7

Everything in post #6 holds. The case it does not cover is the one I have.

Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers).

One of those cases where knowing the mechanism does not help the decision.

5 likes 19mo
NI
n.ibarraTL221 Dec 2024#8
h.iyer, post #6: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. It took me longer than it should have to see that. Go to post

Adding a note of thanks rather than an opinion. I did not know most of that.

0 likes in reply to #6 19mo
QZ
q.zhao_qaTL3Quality assurance22 Dec 2024#9
n.okwuosa, post #4: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Coming back to post #6, because the follow-up matters more than the original answer.

Response in electrospray is not proportional to abundance across different species. Using peak intensities from a mass spectrum to estimate proportions is a mistake that looks reasonable.

2 likes in reply to #4 19mo
ES
e.steinerTL224 Dec 2024#10
p.iyer_pharmd, post #3: Answering the question the opening post raises rather than the one it answers. Mass error in parts per million is (observed minus theoretical) divided by theoretical, times a million. On a high-resolution instrument a low single-digit figure is unremarkable and expected. Go to post

Monoisotopic and average mass are different numbers and both appear on documents. At this molecular weight the difference is a few daltons, which is more than the tolerance being claimed.

0 likes in reply to #3 19mo
GP
g.pemberton_ukTL3Regional · UK25 Dec 2024#11

Picking up post #10: that is the part I would want checked first.

The thing about mass spectrometry that took me longest to accept is that a plausible mechanism is not evidence of an effect. It is a reason to look, not a result.

30 likes 19mo
AV
ai.vukovicTL227 Dec 2024#12

Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied.

I would rather post the uncertainty than round it away.

0 likes 19mo
CR
crossover_reviewTL3Regular28 Dec 2024#13
k.otieno_stats, post #7: Everything in post #6 holds. The case it does not cover is the one I have. Purity and identity are different questions: LC-MS establishes that the species at a retention time has the expected mass. It does not establish how much of the sample is that species (that is what LC-UV purity answers). One of those cases where knowing the… Go to post

For a compound with no reference standard in circulation, an observed mass is much more useful than an assertion of agreement with a theoretical value nobody can check.

5 likes in reply to #7 19mo
KB
k.batistaTL229 Dec 2024#14
e.steiner, post #10: Monoisotopic and average mass are different numbers and both appear on documents. At this molecular weight the difference is a few daltons, which is more than the tolerance being claimed. Go to post

I had written a reply contradicting post #12 and deleted it. Here is what survived.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

15 likes in reply to #10 19mo
MM
methods_marginTL3Regular31 Dec 2024#15

A mass match establishes that the measured mass is consistent with the proposed composition. It does not establish purity, sequence order, or the absence of an isomer, and all three are frequently claimed from it.

Not a conclusion. A place to stand while looking for one.

22 likes 19mo
MG
m.guerreroTL21 Jan 2025 · edited#16

Deamidation adds approximately one dalton and produces a species that frequently elutes very close to the parent. It is the hardest common impurity to see chromatographically and the easiest to see by mass.

If anyone has run this properly I would rather read that than my own guess.

0 likes 19mo
ST
stopper_traceTL22 Jan 2025#17
NH
n.hartmannTL23 Jan 2025#18
methods_margin, post #15: A mass match establishes that the measured mass is consistent with the proposed composition. It does not establish purity, sequence order, or the absence of an isomer, and all three are frequently claimed from it. Not a conclusion. A place to stand while looking for one. Go to post

Coming back to post #16, because the follow-up matters more than the original answer.

Mass accuracy is expressed in parts per million. It is the difference between observed and theoretical mass divided by theoretical mass, multiplied by a million. A high-resolution instrument in good calibration achieves low single-digit ppm on a peptide of this size.

I have written this out at length because the short version keeps being misread.

10 likes in reply to #15 19mo
VS
vial_slopeTL3Regular5 Jan 2025#19

Common adducts: sodium adds ≈22, potassium adds ≈38 compared to hydrogen. A [M+Na]+ peak is common and its mass is predictable from the base mass.

If that reads as pedantic, it is, and it has saved me twice.

0 likes 19mo
MA
m.amankwahTL26 Jan 2025#20
e.steiner, post #10: Monoisotopic and average mass are different numbers and both appear on documents. At this molecular weight the difference is a few daltons, which is more than the tolerance being claimed. Go to post

Agreed, and I will stop repeating the version of this I had been repeating.

2 likes in reply to #10 19mo
MA
m.almeidaTL27 Jan 2025#21

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

That distinction has done more work for me than anything else in this category.

0 likes 19mo
AK
a.kowalczykTL2Regular8 Jan 2025#22
m.almeida, post #21: Resolution: "high resolution" commonly means That distinction has done more work for me than anything else in this category. Go to post

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

The disagreement above is smaller than it looks once the terms are fixed.

21 likes in reply to #21 19mo
VB
v.bruunTL29 Jan 2025#23

Helpful, and easy to find again, which is half of what a good reply is.

9 likes 19mo
KB
k.brandl_deTL311 Jan 2025#24
VK
v.kjaerTL212 Jan 2025#25
p.iyer_pharmd, post #3: Answering the question the opening post raises rather than the one it answers. Mass error in parts per million is (observed minus theoretical) divided by theoretical, times a million. On a high-resolution instrument a low single-digit figure is unremarkable and expected. Go to post

Sample preparation for mass spectrometry can itself introduce modifications, particularly oxidation. A finding at trace level may be telling you about the preparation rather than the material.

29 likes in reply to #3 18mo
DB
d.bramleyTL3Regular13 Jan 2025#26
v.bruun, post #23: Helpful, and easy to find again, which is half of what a good reply is. Go to post

Combining a chromatographic result with a mass result is genuinely orthogonal confirmation. Either alone leaves a specific class of problem invisible, and the two classes do not overlap much.

15 likes in reply to #23 18mo
HJ
h.jansenTL214 Jan 2025#27

Second-hand on mass spectrometry, so weight it accordingly — someone whose method I trust told me this and I have not verified it myself.

5 likes 18mo
G
GEldridgeTL3Regular15 Jan 2025#28

Post #25 is right about the mechanism and I think understates the practical bit.

Offering a way to settle mass spectrometry rather than another opinion about it. Two measurements, taken the same way, a fortnight apart. If the difference is within the noise, the question was not answerable at this precision.

0 likes 18mo
AP
a.pereiraTL216 Jan 2025#29
ai.vukovic, post #12: Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied. I would rather post the uncertainty than round it away. Go to post

On post #25 — agreed on the reasoning, with one qualification.

Trifluoroacetate adducts are common in material purified with TFA and are one reason a mass spectrum from a peptide can look busier than expected.

The part I am sure of is shorter than the part I have written.

22 likes in reply to #12 18mo
PE
ppm_errorTL3Analytical chemist17 Jan 2025#30
h.iyer, post #6: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. It took me longer than it should have to see that. Go to post

Fair, and the limits you put on it are the part I will remember.

10 likes in reply to #6 18mo