The Peptide CommonsEst. May 2024
Independent. We sell nothing and are affiliated with no manufacturer or pharmacy. Every moderation action is logged in public
Analytics · Method validation

An inter-laboratory comparison on one lot, tabulated — what changed since

IA
i.almeidaTL217 Sep 2025#1

An inter-laboratory comparison on one lot, tabulated — what changed since — setting out what I have, and where I think it stops being reliable.

What changes if the standard account of inter-laboratory comparison is wrong? I ask because I have been treating it as settled and I noticed this week that I could not say why.

Working through the consequences rather than the evidence, since others here are better placed on the evidence.

12 likes 10mo
D
DSakamotoTL3Regular25 Sep 2025#2

Why two laboratories may disagree: after validating the same method, they may still report different purity on the same sample due to integration differences, column age differences, subtle differences in mobile phase pH or temperature. This is normal and not a sign that one is wrong.

A qualification I should have led with rather than closed on.

1 like 10mo
CK
c.kuuselaTL230 Sep 2025#3

The arithmetic on inter-laboratory comparison is the easy part and it is where the errors are, which is an uncomfortable combination. Show your working and someone will catch it.

0 likes 10mo
TF
taper_fileTL3Regular4 Oct 2025#4

The arithmetic in post #3 is right; the assumption feeding it is the part to check.

My position on inter-laboratory comparison is current rather than settled. I have revised it once already and I expect to again, so treat it accordingly.

23 likes 10mo
HK
h.krastevTL29 Oct 2025#5

Range and working range are different things and a certificate rarely distinguishes them. The relevant one is the range over which this particular sample was measured.

I have separated what I observed from what I concluded, which does not always happen.

10 likes 10mo
D
DKwiatkowskiTL3Regular13 Oct 2025#6

Validation is compound-specific and matrix-specific. A method validated for one peptide is a starting point for another and not a validated method for it.

That is the version I would defend. It is not the version I started with.

3 likes 9mo
DA
d.achebeTL217 Oct 2025#7

Everything in post #3 holds. The case it does not cover is the one I have.

The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number.

0 likes 9mo
N
NorringtonTL3Regular20 Oct 2025 · edited#8
i.almeida, post #1: An inter-laboratory comparison on one lot, tabulated — what changed since — setting out what I have, and where I think it stops being reliable. What changes if the standard account of inter-laboratory comparison is wrong? I ask because I have been treating it as settled and I noticed this week that I could not say why. Working through… Go to post

Narrowing post #7, because the general version has more than one answer.

I read the earlier replies on inter-laboratory comparison twice before writing this, because I had assumed the opposite and wanted to be sure I was disagreeing with what was said rather than what I expected.

31 likes in reply to #1 9mo
SO
sa.okonkwoTL224 Oct 2025#9

I had written a reply contradicting post #7 and deleted it. Here is what survived.

Inter-laboratory comparison has a well-known answer and a correct answer, and the interesting work is establishing that they are the same. Nobody has done that here yet.

16 likes 9mo
JD
j.delacroixTL3Regular27 Oct 2025#10

Confirming post #7 from a second method, which matters more than confirming it from a second person.

The most useful single question about a method: what would it fail to detect? Every method has an answer and few documents state it.

6 likes 9mo
EK
e.kuuselaTL231 Oct 2025#11

What I would want before treating inter-laboratory comparison as settled: the method, the sample, and whether anyone tried to find the opposite result. Two of the three are usually missing.

0 likes 9mo
C
chromatogramTL4Analytical chemist3 Nov 2025#12

Reading back through the inter-laboratory comparison threads from last year, the same three questions come up every time and only one of them has ever been answered properly. That seems like a documentation gap rather than a knowledge gap.

0 likes 9mo
SC
s.coelhoTL27 Nov 2025#13
c.kuusela, post #3: The arithmetic on inter-laboratory comparison is the easy part and it is where the errors are, which is an uncomfortable combination. Show your working and someone will catch it. Go to post

Confirming post #12 from a second method, which matters more than confirming it from a second person.

Where a pharmacopoeial monograph exists, a method that follows it inherits a great deal of assurance. Almost nothing discussed here has one.

The answer changed when I changed how I was measuring, which was informative.

5 likes in reply to #3 9mo
JM
j.mwangiTL4 Moderator10 Nov 2025#14

I had written a reply contradicting post #10 and deleted it. Here is what survived.

The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number.

That is one dataset and I would not build a rule on it.

14 likes 9mo
RP
r.petrovTL213 Nov 2025#15
P
preregisteredTL3Research methods16 Nov 2025#16

Marking my uncertainty on inter-laboratory comparison explicitly. I am confident about the direction, much less confident about the size, and not confident at all that it generalises past the case in the first post.

2 likes 8mo
MA
m.adeyemiTL219 Nov 2025#17
c.kuusela, post #3: The arithmetic on inter-laboratory comparison is the easy part and it is where the errors are, which is an uncomfortable combination. Show your working and someone will catch it. Go to post

Stability-indicating method: one that can separate a compound from its degradation products. Critical for assay methods that claim to measure actual degradation (as opposed to purity, which is orthogonal).

9 likes in reply to #3 8mo
RI
retention_indexTL2Analytical chemist22 Nov 2025#18

Where I part company with post #14, and it is a narrow parting.

An independent laboratory's method being different from the supplier's is not a discrepancy. It becomes one only when the results differ by more than both methods' demonstrated precision.

That is the version I use. It may not be the version that is correct.

20 likes 8mo
AK
a.kirchnerTL225 Nov 2025#19

Narrowing post #16, because the general version has more than one answer.

Transfer between laboratories: a method can be transferred from one lab to another, but the receiving lab needs to demonstrate that they can achieve the same performance. This requires comparative testing and sometimes small method refinements.

I would put the burden of proof on the interesting explanation, not the dull one.

0 likes 8mo
LI
l.ibarraTL2Regular28 Nov 2025#20

Two people in this thread mean different things by inter-laboratory comparison and are disagreeing about the definition while believing they are disagreeing about the facts. Worth pausing to define it.

5 likes 8mo
OF
outline_firstTL3Wiki editor1 Dec 2025#21

Post #19 and I disagree about the size of the effect, not about the direction.

A stability-indicating method is one demonstrated to resolve the analyte from its degradation products, usually through forced degradation. Calling a method stability-indicating without that work is a claim rather than a property.

22 likes 8mo
CC
c.castellanosTL24 Dec 2025#22

System suitability is the ongoing evidence that a validated method is still performing. Validation is done once; suitability is done every run, and it is the one that appears on a certificate.

Noting that I have skin in this question and have tried to discount for it.

10 likes 8mo
NE
n.ekstromTL2Regular7 Dec 2025 · edited#23
h.krastev, post #5: Range and working range are different things and a certificate rarely distinguishes them. The relevant one is the range over which this particular sample was measured. I have separated what I observed from what I concluded, which does not always happen. Go to post

Adding a small correction to the inter-laboratory comparison summary above rather than a disagreement with it. The substance holds; one of the figures is out by a factor that matters.

1 like in reply to #5 8mo
SV
s.vukovicTL210 Dec 2025#24
n.ekstrom, post #23: Adding a small correction to the inter-laboratory comparison summary above rather than a disagreement with it. The substance holds; one of the figures is out by a factor that matters. Go to post

Narrowing post #23, because the general version has more than one answer.

Inter-laboratory comparison is a good example of a question where the honest answer is boring and the interesting answers are unsupported. I would go with boring.

0 likes in reply to #23 8mo
SS
steady_stateTL3Regular12 Dec 2025#25

Worth separating two things that post #23 runs together.

A method that reports the same number for every lot is worth a second look. Real processes vary, and a total absence of variation is a statement about the measurement rather than the process.

16 likes 7mo
NC
n.cabreraTL215 Dec 2025#26

That is a fair summary of where the discussion has got to.

6 likes 7mo
JW
journalclub_wrenTL3Regular18 Dec 2025#27
c.kuusela, post #3: The arithmetic on inter-laboratory comparison is the easy part and it is where the errors are, which is an uncomfortable combination. Show your working and someone will catch it. Go to post

The most useful single question about a method: what would it fail to detect? Every method has an answer and few documents state it.

I am reporting what happened, not recommending it.

0 likes in reply to #3 7mo
EH
e.halonenTL221 Dec 2025#28

Whatever the answer on inter-laboratory comparison turns out to be, the method for getting there is the same: state the assumption, do the arithmetic in public, invite the correction.

31 likes 7mo
DH
dietitian_hollisTL3Dietitian23 Dec 2025#29

The limit of quantitation determines what the impurity table can honestly contain. Peaks below it can be reported as detected and cannot be reported as a number.

Take it as a starting point and not as a specification.

0 likes 7mo

Suggested topics

TopicParticipantsRepliesViewsActivity
About the Method validation category
Specificity, linearity, precision, accuracy, robustness, system suitability, and why two labs disagree. This post is a community wiki: any member at trust level 3 or above can edit it, and every edit is…
KTVPAKAJV+1 5 468 5mo
Sample handling between despatch and injection
Sample handling between despatch and injection Writing it up because I had to work it out twice and would rather nobody else did. Trying to work out what would count as evidence on sample handling between…
KREHSSNCNA+63 68 2.4k 3mo
LOD and LOQ: definitions people get wrong
LOD and LOQ: definitions people get wrong — setting out what I have, and where I think it stops being reliable. What changes if the standard account of LOD and LOQ is wrong? I ask because I have been treating…
EISFCN 2 43k 8mo
Revisiting: Sample handling between despatch and injection
Posting this under the heading it deserves: Revisiting: Sample handling between despatch and injection Everything below is what sits behind that. Reading back through what has been written here about Sample…
DVEAGTEMNG+5 9 1.1k 6mo
Validating a method you did not develop
On the subject in the title: Validating a method you did not develop Working notes rather than a conclusion. What changes if the standard account of validating a method is wrong? I ask because I have been…
CRSOSSSVJW+10 14 15k 3mo

Related topics — sharing the tags stability-indicating method, LOD & LOQ, system suitability

TopicParticipantsRepliesViewsActivity
Validating a method you did not develop
On the subject in the title: Validating a method you did not develop Working notes rather than a conclusion. What changes if the standard account of validating a method is wrong? I ask because I have been…
CRSOSSSVJW+10 14 15k 3mo
Shoulders, tangent skims, and defensible integration — the long version
Shoulders, tangent skims, and defensible integration — the long version Writing it up because I had to work it out twice and would rather nobody else did. Setting out an integration problem with the actual…
VSCBAWSLEL+16 20 27k 3mo
What consumer-accessible tests can and cannot determine — a second dataset
What consumer-accessible tests can and cannot determine — a second dataset I have a specific reason for asking rather than idle curiosity, and the context is below. Posting a small dataset on…
IZBNBPNN+117 131 45k 10h
Relative retention time as a better inter-laboratory comparator
On the subject in the title: Relative retention time as a better inter-laboratory comparator Working notes rather than a conclusion. What changes if the standard account of relative retention time is wrong? I…
MCVKSKHVCL+7 11 28k 14mo
Coming back to: Where the integrator drew the baseline, and how much it moved the number
On the subject in the title: Where the integrator drew the baseline, and how much it moved the number Working notes rather than a conclusion. Posting the method first, because I know what the first three…
AHNMTNAFMB+20 24 2.7k 10mo