a.cabreraTL22 Feb 2026#1

Posting this under the heading it deserves: MS/MS sequence confirmation: what it costs and when it is worth it — what changed since Everything below is what sits behind that.

Two reports on the same lot, and I would like help deciding whether they disagree.

The first gives 98.6% by reversed-phase at 220 nm. The second gives 97.6% by a method described only as HPLC. Neither states the integration convention.

My instinct is that this gap is method rather than material. I would like that instinct checked by somebody who does this for a living rather than confirmed by somebody who agrees with me.

16 likes 6mo
CC
crossref_checkTL3Wiki editor10 Feb 2026 · edited#2

I read the opening post twice before replying, because I had assumed the opposite.

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

Written in the hope of being told what I have missed.

20 likes 6mo
ND
n.duarteTL216 Feb 2026#3
a.cabrera, post #1: Posting this under the heading it deserves: MS/MS sequence confirmation: what it costs and when it is worth it — what changed since Everything below is what sits behind that. Two reports on the same lot, and I would like help deciding whether they disagree. The first gives 98.6% by reversed-phase at 220 nm. The second gives 97.6% by a… Go to post

Isotope pattern is a check that costs nothing to look at. A species at this molecular weight has a characteristic envelope, and something with the wrong envelope is not what it says it is.

Not the whole picture, but the part of it I can speak to.

0 likes in reply to #1 5mo
V
VPoulsenTL3Regular21 Feb 2026#4

Noted, and I have changed what I was going to do on the strength of it.

2 likes 5mo
VB
v.bergstromTL226 Feb 2026#5

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

5 likes 5mo
RJ
r.jhannsdttirTL3Regular2 Mar 2026#6

Everything in post #2 holds. The case it does not cover is the one I have.

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

That matches what I was told, which is not the same as knowing it.

14 likes 5mo
AV
a.vermeulenTL26 Mar 2026#7
crossref_check, post #2: I read the opening post twice before replying, because I had assumed the opposite. Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Written in the hope of being told what… Go to post

Taking post #5 at face value and following it one step further.

Electrospray on a peptide of this size gives a multiply charged series rather than a single ion. Seeing only one charge state usually means the deconvolution has already been done for you, which is worth knowing.

Adding the caveat now so it does not have to be extracted later.

0 likes in reply to #2 5mo
TK
t.kulkarniTL3Regular10 Mar 2026#8

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

Noting that I have skin in this question and have tried to discount for it.

0 likes 5mo
GO
g.oyelaranTL214 Mar 2026#9

What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else.

19 likes 4mo
HA
h.almeidaTL2Member18 Mar 2026#10
n.duarte, post #3: Isotope pattern is a check that costs nothing to look at. A species at this molecular weight has a characteristic envelope, and something with the wrong envelope is not what it says it is. Not the whole picture, but the part of it I can speak to. Go to post

A mass match establishes that the measured mass is consistent with the proposed composition. It does not establish purity, sequence order, or the absence of an isomer, and all three are frequently claimed from it.

0 likes in reply to #3 4mo
B
BBramleyTL3Regular22 Mar 2026#11
VPoulsen, post #4: Noted, and I have changed what I was going to do on the strength of it. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

Marking that as an opinion rather than a finding.

6 likes in reply to #4 4mo
CS
c.serranoTL226 Mar 2026#12

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

I would put the burden of proof on the interesting explanation, not the dull one.

1 like 4mo
TN
t.nardoneTL3Regular29 Mar 2026#13

Where I part company with post #9, and it is a narrow parting.

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

0 likes 4mo
IA
i.amankwahTL22 Apr 2026#14
g.oyelaran, post #9: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

22 likes in reply to #9 4mo
LC
l.chevalierTL3Regular5 Apr 2026#15
t.nardone, post #13: Where I part company with post #9, and it is a narrow parting. Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible. Go to post

Calibration state at the time of the run determines whether the ppm figure means anything. A report that states when the instrument was last calibrated is unusual and is worth more than one that does not.

3 likes in reply to #13 4mo
EM
e.mensaTL29 Apr 2026#16

Reading rather than contributing, but this is the most useful thread I have found on it.

0 likes 4mo
VD
vial_deskTL3Regular12 Apr 2026#17

I had written a reply contradicting post #13 and deleted it. Here is what survived.

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

That is one dataset and I would not build a rule on it.

31 likes 4mo
AE
a.eriksenTL216 Apr 2026 · edited#18

Confirming post #17 from a second method, which matters more than confirming it from a second person.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

16 likes 3mo
EC
excursion_checkTL3Regular19 Apr 2026#19
l.chevalier, post #15: Calibration state at the time of the run determines whether the ppm figure means anything. A report that states when the instrument was last calibrated is unusual and is worth more than one that does not. Go to post

I read post #17 twice before replying, because I had assumed the opposite.

Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied.

1 like in reply to #15 3mo
SV
s.vanheckeTL222 Apr 2026#20
BBramley, post #11: Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation. Marking that as an opinion rather than a finding. Go to post

Combining a chromatographic result with a mass result is genuinely orthogonal confirmation. Either alone leaves a specific class of problem invisible, and the two classes do not overlap much.

0 likes in reply to #11 3mo
MP
mira.patelTL4 Admin25 Apr 2026#21

Electrospray on a peptide of this size gives a multiply charged series rather than a single ion. Seeing only one charge state usually means the deconvolution has already been done for you, which is worth knowing.

Anyone with a larger sample, please post it.

3 likes 3mo
EV
e.vargaTL228 Apr 2026#22
h.almeida, post #10: A mass match establishes that the measured mass is consistent with the proposed composition. It does not establish purity, sequence order, or the absence of an isomer, and all three are frequently claimed from it. Go to post

Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.

11 likes in reply to #10 3mo
AA
a.adebayoTL22 May 2026 · edited#23
vial_desk, post #17: I had written a reply contradicting post #13 and deleted it. Here is what survived. Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory. That is one dataset and I would not build a rule on it. Go to post

Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.

32 likes in reply to #17 3mo
AN
a.novakTL25 May 2026#24
DT
d.tammTL28 May 2026#25

I came in to disagree and I am leaving without a disagreement.

1 like 3mo
BF
b.fonsecaTL211 May 2026#26
g.oyelaran, post #9: What mass accuracy establishes: the measured mass is consistent with a specific composition. What it does not establish: purity, sequence order, stereochemistry, or the absence of an isobaric species. Every one of those requires something else. Go to post

Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.

The right answer here may simply be that it has not been measured.

7 likes in reply to #9 3mo
CS
c.silvaTL214 May 2026#27

Tandem mass spectrometry: MS/MS fragments the molecular ion and uses fragment masses to confirm identity and detect modifications. A simple identity confirmation by LC-MS does not address post-translational modifications or impurities with the same or very close mass.

24 likes 2mo
AZ
a.zamoraTL217 May 2026#28

A mass match establishes that the measured mass is consistent with the proposed composition. It does not establish purity, sequence order, or the absence of an isomer, and all three are frequently claimed from it.

0 likes 2mo
EF
e.ferreiraTL3Regular20 May 2026#29

Picking up post #26: that is the part I would want checked first.

The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.

Happy to expand any of that if it is the useful part.

10 likes 2mo
AJ
a.jansenTL223 May 2026#30

On post #28 — agreed on the reasoning, with one qualification.

Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.

Scoping that to what I have actually seen rather than what I have read.

23 likes 2mo