Quantitation by MS: most quantitation is done by LC-UV detection at 214 nm, not by MS, because extinction coefficients are better known. MS can quantify if an internal standard is used but that requires preparation.
MS/MS sequence confirmation: what it costs and when it is worth it — what changed since posts 31–40
This is a continuation of a long topic, addressed by post number rather than by page. Start at post 1.
Positional isomers and epimers are mass-identical. Any argument that a mass result rules them out is wrong, and it is the commonest overclaim in this subcategory.
That much is documented. The rest is how I have interpreted it.
That is consistent with mine, for whatever one more account is worth.
Post #31 is the version of this I will quote in future. One addition.
Resolution and mass accuracy are different specifications. An instrument can resolve two species and still assign their masses imprecisely, and the reverse is also possible.
The conclusion is tentative; the arithmetic underneath it is not.
The honest summary of what a mass result buys you: it narrows the field of what the material could be, considerably. It never closes it, and no certificate should be read as though it had.
Nothing above should be read as advice about what anyone else should do.
Picking up post #35: that is the part I would want checked first.
Tandem mass spectrometry with fragmentation gives sequence information that intact mass cannot. It is the analysis that would actually confirm identity, and it is rarely supplied.
That is all I can say without guessing.
I had written a reply contradicting post #35 and deleted it. Here is what survived.
Resolution: "high resolution" commonly means <5 ppm across the mass range. Unit-resolution instruments achieve ±1 Da at best and cannot distinguish two species differing by less than 1 Da in total mass.
Worth separating two things that post #35 runs together.
Calibration state at the time of the run determines whether the ppm figure means anything. A report that states when the instrument was last calibrated is unusual and is worth more than one that does not.
That is what the documentation says. What happens in practice is usually close.
Thank you for the correction. I would rather find out here than later.
This topic was referenced in
- MS/MS sequence confirmation: what it costs and when it is worth itAnalytics › Mass spectrometry · 6 replies
Suggested topics
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Why mass spectrometry cannot tell you about an epimer
Why mass spectrometry cannot tell you about an epimer I have a specific reason for asking rather than idle curiosity, and the context is below. I have spent a fortnight trying to pin mass spectrometry down…
|
+75 | 82 | 4k | 17mo |
|
Distinguishing a deletion sequence by mass alone
Distinguishing a deletion sequence by mass alone Writing it up because I had to work it out twice and would rather nobody else did. Trying to establish the boundary conditions on distinguishing a deletion…
|
+121 | 129 | 25k | 17mo |
|
What a 2 ppm match establishes, and what it does not
What a 2 ppm match establishes, and what it does not — that is the question, and I have not found it answered plainly anywhere I have looked. Working through the identity arithmetic and I would like it…
|
+56 | 60 | 5.2k | 9mo |
|
High-resolution MS on aggregates and what it can see — one year on
High-resolution MS on aggregates and what it can see — one year on Writing it up because I had to work it out twice and would rather nobody else did. A narrow question about High-resolution, deliberately…
|
+54 | 58 | 1.2k | 23mo |
|
Distinguishing a deletion sequence by mass alone — the long version
On the subject in the title: Distinguishing a deletion sequence by mass alone — the long version Working notes rather than a conclusion. A narrow question about Distinguishing a deletion sequence,…
|
+38 | 44 | 60k | 11mo |
Related topics — sharing the tags charge states, LC-MS, MALDI-TOF
| Topic | Participants | Replies | Views | Activity |
|---|---|---|---|---|
|
Coming back to: Oxidation and deamidation mass shifts, tabulated
Oxidation and deamidation mass shifts, tabulated — setting out what I have, and where I think it stops being reliable. Reading back through what has been written here about Oxidation and deamidation mass…
|
+34 | 38 | 4.8k | 10mo |
|
Second pass at: What a 2 ppm match establishes, and what it does not
Second pass at: What a 2 ppm match establishes, and what it does not — setting out what I have, and where I think it stops being reliable. Two things I would like separated before anyone answers on 2 ppm…
|
+44 | 49 | 8.1k | 1d |
|
Charge states for a 4 kDa peptide, worked through — the long version
Charge states for a 4 kDa peptide, worked through — the long version Writing it up because I had to work it out twice and would rather nobody else did. Working through the identity arithmetic and I would like…
|
2 | 293 | 3mo | |
|
Endotoxin and why it is a separate question from purity — what changed since
On the subject in the title: Endotoxin and why it is a separate question from purity — what changed since Working notes rather than a conclusion. What changes if the standard account of Endotoxin is wrong? I…
|
+78 | 91 | 1k | 10h |
|
Isotonicity and what happens when you get it wrong
On the subject in the title: Isotonicity and what happens when you get it wrong Working notes rather than a conclusion. I was wrong about isotonicity in a thread last spring and I would like to correct it…
|
+27 | 31 | 11k | 18mo |